Hello,
I was aware of the post a few days ago, but as you have opened the new thread, I comment here.I carry little with the bomb, just a couple of months, and I had the same problem.Modify the procedure that followed after reading the post that says @fer.I summarize how I do it:
1. After opening the reservoir, I move the embolus down and up to turn it so that the fat that the two gum joints of the piston have extends through the inside of the reservoir.
2. I connect the reservoir with the insulin pen and I load the reservoir slowly 30 in 30 units.(Before towards the change of reservoir in the morning and I think that in doing so with hurry and loading the reservoir quickly, bubbles were made).
3. In the penultimate load of the reservoir, I give some tap to unify the bubbles that have been created.Then I pushed the reservoir piston to pass the bubbles and some insulin to the pen.You have to do some strength on the piston, once the bubbles have passed to the pen we stop pushing the reservoir embolo.With this we managed to remove bubbles without wasting insulin.I suppose it is obvious, but all these steps are taken with the pen vertically and the reservoir underneath.
4. After loading the insulin I remove the pen and hit the reservoir so that the bubbles that may remain go up.I purge the bubbles that may remain.
5. With the reservoir down I Quito the adapter.
6. I connect the cannula cable to the reservoir.Yes there are still bubbles can be purged manually.
7. I make the set change following the instructions of the pump.Here, the pump always vertically and bumps (the gross) while raising the piston to the reservoir and while the tube is filled.In this step many times I see microburbujas to the tube.Stop once is the tube full of insulin without bubbles.
There are also several things about the infusion set that I wanted to comment (I use the quick-set paradigm, for the photos it seems that the reservoir is not the same as I have).
With the passage of time in the tube, white areas appear in the plastic.These areas may seem bubble but they are not.
Do not lose sight of when we see bubbles is the amount of insulin to which they are equivalent.To the gross, if the tube enter 6 units and a bubble of 1 cm has a length of 60 cm corresponds to 0.1 ud of insulin.
Another point with which I found is that in case of bubbles during use I did not know how to purge them.What I do is disconnect the body tube , remove the reservoir and do as if we made a set change.When filling the tube, we fill it until the bubbles go and when we finish we do not fill the cannula (it is already full).
Well, this is what I do.If someone is delighted.
Greetings.